| Hello, I do denaturing SDS polyacrylamide gel electrophoresis (SDS PAGE) according to Laemmli (i.e. with a stacking gel and a separating gel) for analytical separation of proteins. I use a vertical BioRad gel electrophoresis apparatus. 1) One of my problems is that the samples tend to run towards the middle, i.e. they do not run in a straight vertical line but rather bent towards the middle. This effect is strongest for the outer lanes. 2) Another problem is that my bands become diffuse in the lower molecular weight region although I use a 15 % SDS polyacrylamide gel. The weight of proteins in that region ranges from about 18 kDa to 30 kDa. Literature says proteins of that size can still be separated by SDS PAGE. What could these effects be due to? What am I doing wrong? What can I do to get prettier results in my SDS PAGE? Peter |
|
#2
| |||
| |||
| "Peter Frank" <peter_frankde@yahoo.de> wrote in message news This problem usually arises when the samples contain different salt concentrations. Equalize the concentration of salt in all samples and if you are not using all the lanes, load the empty lanes with SDS-PAGE load buffer containing the same salt concentration as your sample lanes. |
|
#3
| |||
| |||
| I regularly used a Laemmli system with additional urea, and good flushing of the sample wells before loading was quite important in order to avoid this problem. Besides, I did a prerun of about 10-15 min before loading the samples, although this is not recommended for this (discontinue) system. Oh well, whatever works! Rene. "Kyle Legate" <[Only registered users see links. ]> wrote in message news:<a75c1648c1a3915cae66f4d2ba50bcbf@news.terane ws.com>... |
|
#4
| |||
| |||
| Kyle is right. It is rather frustrating at times when the sample volumes differ greatly due to the protein concentrations. I have, in my life so far, run close to a thousand protein gels of different kinds and a few important things to have a "pretty" picture that I have settled on are: 1. Very clean plates 2. Very clean, dust-free spacers (especially at the edge where they make contact with the gel. If using alcohol to wipe the components, make sure no trace of that is left. 3. As-close-to-equal volume loads as possible with low salt concentration. Salt may be dialyzed out using microdialysis on a GVWP membrane (takes about 20 minutes) 4. Separating gel polymerization time should not be less than an hour. 5. The Stacking layer (low pH) should not be poured until one is ready with the samples and is absolutely sure that the electrophorsis will begin within about an hour! 6. The wells should be cleaned with the tank/running buffer using a syringe prior to loading samples. 7. The buffer components used (including the running buffer) should be of good quality. 8. Although most protocols talk about constant current, constant voltage works much better in producing "pretty" pictures. The decreasing PD under constant current causes the low Mr polypeptides to show up diffused. Also, the voltage/current should be kept at the optimum. 9. The volume of the load should preferably be kept low. Thumb rule: the height of the load less than half of the stacking gel height. 10. for a regular sized (20cm) 1.5mm gel, no more than 100ug of a total crude extract protein/lane. For 0.75mm, it should be <50ug/lane. For partially purified proteins, it is <50ug/lane and <20ug/lane. Then again, these values are better impiricaly determined. Quoting Kyle Legate <[Only registered users see links. ]>: -- Hiranya S. Roychowdhury, Ph.D. Coll. Asst. Professor, Molecular Biology, Dept. of Chemistry & Biochemistry Rm# 336, Chemistry Bldg.; MSC 3MLS New Mexico State University Las Cruces, NM 88003 --- |
|
#5
| |||
| |||
| "Dr. Hiranya S. Roychowdhury" wrote: I agree with all of these points, with two comments: I have routinely poured 30 gels at a time in a multicaster and stored them upright, combs intact, in a sealable plastic container with a thin layer of water at the bottom, in the coldroom for 2-3 weeks at a time with no noticeable decrease in quality. I run Biorad mini gels (~5cm separating length) at 200 volts constant voltage. It takes around 45 minutes for the dye to reach the bottom, and I have never had a problem with fuzzy bands above, say, 6 kDa. |
|
#6
| |||
| |||
| Thanks for the many useful tips! Peter Dr. Hiranya S. Roychowdhury wrote: |
|
#7
| |||
| |||
| I haven't had such luck with even when I pourted the stacking gel the night before. And, I did mention that constant voltage is superior to constant current so far as "crispy" separation is concerned. For me, mini gels (0.75mm) run in 45 min at 150V :-) Quoting Kyle Legate <[Only registered users see links. ]>: -- Hiranya S. Roychowdhury, Ph.D. Coll. Asst. Professor, Molecular Biology, Dept. of Chemistry & Biochemistry Rm# 336, Chemistry Bldg.; MSC 3MLS New Mexico State University Las Cruces, NM 88003 --- |
|
#8
| |||
| |||
| Peter Frank wrote: 1) This could be an effect of gel heating. Make sure that the current is not too high (10-15 mA/gel should be all right), salt concentration in the sample is ALARA and that you use the cooling loop of your apparatus. Best cooling is achieved with a cooling water bath set to about 15 degrees C. 2) Try gradient gels for crisper bands. For a 18-30 kDa weight range I would probably try a 10-15% T gradient. 5-15% if you need higher MW proteins as well. |
|
#9
| |||
| |||
| On Fri, 19 Dec 2003, Dr Engelbert Buxbaum wrote: The description certainly sounds like the phenomenon known as 'gel smiling', which is indeed due to overheating. tom -- Tom Anderson, MRC Laboratory for Molecular Cell Biology, UCL |
|
#10
| |||
| |||
| On Fri, 19 Dec 2003, Tom Anderson wrote: Actually, now i read it again more closely, it doesn't at all. Sorry. tom -- Tom Anderson, MRC Laboratory for Molecular Cell Biology, UCL |
| Tags |
| electrophoresis , gel , middle , running , samples |
| Thread Tools | |
| Display Modes | |
|
|
| | ||||
| Thread | Thread Starter | Forum | Replies | Last Post |
| USA - custom petides? | avinash | Proteomics Forum | 6 | 06-09-2010 04:10 AM |
| SDS-PAGE Protocol | moleculardude | Protein Science | 14 | 02-12-2010 10:27 AM |
| Dot instead of a band on precast gel (running funny!) | aditi | Western Blot Forum | 3 | 10-15-2009 09:24 PM |
| loading RNA samples on denat. PAGE | Simone Marker | Protocols and Methods Forum | 0 | 01-21-2008 09:55 AM |