PFGE troubleshooting

Molecular Biology Techniques

Molecular Biology Forum. Includes forums for common molecular biology techniques.



Register Molecular Biology Forums
Reply
 
LinkBack Thread Tools Display Modes
  #1 (permalink)  
Old 03-12-2008, 06:22 PM
Pipette Filler
Points: 63, Level: 1Points: 63, Level: 1Points: 63, Level: 1
Activity: 0%Activity: 0%Activity: 0%
 

Join Date: Mar 2008
Posts: 2
swagemann RSS Feed
Default PFGE troubleshooting

I'm having major problems with our PFGE. We've had Bio-Rad's CHEF DRIII setup for almost 10 years now. Our main project is to compare E. coli from a host animal species and compare it to a water sample to determine if there is any human dna found in the water body.
We were funded for 8 years by various entities, and the funding dried up. After 2 years we were contacted to run some samples. For the past year, I have not gotten any decent results. I've made up the reagents in our protocol 4 or 5 times, and even tried using Bio-Rad 'store bought' chemicals, and nothing seems to work. Here is what we know: we don't think it's the equipment. Lambda ladder control plugs are coming out fine. We don't think it's in the restriction enzyme step. We had an old Bio Rad JM107 E. coli plug (they no longer make it) and used our restriction enzyme onit, and it came out fine. There are times when it just looks like the lane is smeared. Mostly it looks like we get decent bands towards the top of the gel, but the bands then get progressively worse as we go down the gel.
Any one have any ideas?
Thanks,
Scott
Digg this Post!Add Post to del.icio.usBookmark Post in TechnoratiFurl this Post!Spurl this Post!Reddit!
Reply With Quote
Alt Today
Advertising
Google Adsense
 
This advertising will not be shown
in this way to registered members.
Register your free account today
and become a member on
Molecular Biology Forum
Standard Sponsored Links

  #2 (permalink)  
Old 06-14-2008, 09:54 PM
Pipette Filler
Points: 102, Level: 2Points: 102, Level: 2Points: 102, Level: 2
Activity: 0%Activity: 0%Activity: 0%
 

Join Date: May 2008
Posts: 3
chovek69 RSS Feed
Default Re: PFGE troubleshooting

Quote:
Originally Posted by swagemann View Post
I'm having major problems with our PFGE. We've had Bio-Rad's CHEF DRIII setup for almost 10 years now. Our main project is to compare E. coli from a host animal species and compare it to a water sample to determine if there is any human dna found in the water body.
We were funded for 8 years by various entities, and the funding dried up. After 2 years we were contacted to run some samples. For the past year, I have not gotten any decent results. I've made up the reagents in our protocol 4 or 5 times, and even tried using Bio-Rad 'store bought' chemicals, and nothing seems to work. Here is what we know: we don't think it's the equipment. Lambda ladder control plugs are coming out fine. We don't think it's in the restriction enzyme step. We had an old Bio Rad JM107 E. coli plug (they no longer make it) and used our restriction enzyme onit, and it came out fine. There are times when it just looks like the lane is smeared. Mostly it looks like we get decent bands towards the top of the gel, but the bands then get progressively worse as we go down the gel.
Any one have any ideas?
Thanks,
Scott
OK. I cannot say that I am an expert but currently I am dealing with standardization of PFGE protocols for typing various nosocomial pathogens. We use Gene Navigator System (Amersham, now GE).So far I've made two gels (about 30 strains)- the first was fine (except that the time was too much) ...the second not so well, but acceptable. Maybe this is the newbie luck but here are some things I can say for sure. The cell concentration is extremely important.....0,8 to 1 OD measured at 600-610nm seems to work best. Do you have lysozyme step prior to prot K? How many times, at what temp and with what buffer do you wash samples after lysis. I wash 3 times with 2ml/2 blocks per tube TE at 4C, 1/2h each and then 100mcL restriction buffer for 1/2h at 4C; change restric. buf. and then do the restriction in 100mcl fresh buffer.
Do you add thiourea in the TBE? Check the literature- it helps alot...
As you see there are alot of factors contributing to failure ....that's why I truly hate this technique...besides there are so many other more discriminative and far more cheap methods. But all is a matter of economical interests....

Good luck
Digg this Post!Add Post to del.icio.usBookmark Post in TechnoratiFurl this Post!Spurl this Post!Reddit!
Reply With Quote
Reply

Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On
Forum Jump

Similar Threads
Thread Thread Starter Forum Replies Last Post
Western Blot Background Troubleshooting molecule2005 Western Blot Forum 11 07-01-2008 09:24 PM
Sandwish ELISA Problems and Troubleshooting proatpcr ELISA Assay Forum 1 04-13-2008 05:07 PM
DNA Ligation Not Working Troubleshooting gandalfthegrey DNA Techniques 1 09-17-2007 07:14 AM
2-D Gel No Bands Troubleshooting khenwood Proteomics Forum 0 08-27-2007 11:10 AM
DNA troubleshooting felihaha DNA Techniques 0 07-04-2007 08:20 AM


All times are GMT. The time now is 07:55 PM.


Powered by vBulletin® Version 3.7.1
Copyright ©2000 - 2008, Jelsoft Enterprises Ltd.
Copyright 2005-2007 Molecular Station | All Rights Reserved