melting point apparatus ?

melting point apparatus ? - Chemistry Forum

melting point apparatus ? - Chemistry Forum. Discuss chemical reactions, chemistry.


Reply
 
LinkBack Thread Tools Display Modes
  #1  
Old 11-07-2003, 10:54 PM
information_person
Guest
 
Posts: n/a
Default melting point apparatus ?



Hello,

Is there a big difference between the melting point apparatus that is
manual, and digital with respect to the
USP requirements ?

I want to get a great quality and accurate melting point apparatus but they
are all over $2000 USD., the manual ones fitted with a digital meter are
$1200 USD.

Can anyone suggest links or melting point apparatus that I can buy from
anywhere that is high quality and acceptable for USP ? should i get manual
or digital ? anyone with knowledge please advice.



Reply With Quote
  #2  
Old 11-08-2003, 11:42 PM
Mark Tarka
Guest
 
Posts: n/a
Default melting point apparatus ?

"information_person" <[Only registered users see links. ]> wrote in message news:<[Only registered users see links. ]>...

Hello.


Yes. You are very fortunate, as I am
trained to guide you through the complexities.


For $1100 cash USD, in advance, I can
provide a high quality melting-point
device for you, sahib.


You should be reading my earlier words,
sahib. That's $1000 cash in advance :-)


Mark (Bush melts at....)
Reply With Quote
  #3  
Old 11-09-2003, 04:35 PM
William Penrose
Guest
 
Posts: n/a
Default melting point apparatus ?

On 8 Nov 2003 15:42:23 -0800, [Only registered users see links. ] (Mark Tarka) wrote:



Information Person:

Keep Mike talking until he's paying you $100, then take it.

Bill Penrose
Reply With Quote
  #4  
Old 11-09-2003, 08:24 PM
R Molony
Guest
 
Posts: n/a
Default melting point apparatus ?

Look in any good older practical org chem book.
The old aluminium block drilled appropriately with a magnifier and
small bunsen burner is still very satisfactory.
Bob M
Reply With Quote
  #5  
Old 11-10-2003, 02:49 AM
Steve Turner
Guest
 
Posts: n/a
Default melting point apparatus ?

[Only registered users see links. ] (R Molony) wrote:


Well, if you want to go the prehistoric route why not use a Thiele
tube?

Steve Turner

Real address contains worldnet instead of spamnet
Reply With Quote
  #6  
Old 11-10-2003, 06:35 PM
R Molony
Guest
 
Posts: n/a
Default melting point apparatus ?

Steve Turner <[Only registered users see links. ].net> wrote:


Naw They break and drop conc Sulphuric over my nice bench.
Bob M
Reply With Quote
  #7  
Old 11-10-2003, 08:58 PM
Chris R. Lee
Guest
 
Posts: n/a
Default melting point apparatus ?

Basically, the pharmaceutical melting point isn't the same as that used by
chemists and others.

In the days before there was any kind of organic spectroscopy, the melting
point was a very important property of a compound. It isn't exactly a point,
for practical reasons and because impurities tend to spread it out in a way
that depends (in practice, forget the theory) on the rate of heating.

If you're writing a chemistry paper for the primary literature, you should
give a range: the lower temperature is where the product just starts to show
signs of melting, and the upper limit is the point where the last particle
has melted.

In pharmaceutical practice, you record only the upper limit, for a quantity
of product in a so-called capillary tube that seems astronomical in other
contexts. There's nothing wrong with this provided the heating rate is
defined - it's 1C/min if I remember rightly. The objective in this context
is to identify the compound, not in the absolute sense, but with respect to
other compounds it might be mistaken for. The only problem is that some
people of a regulatory bent want a melting point when there isn't one; if
there are any signs of decomposition refuse to quote a value.

To answer you question, you can use a flask of oil heated by a bunsen,
provided you keep within an undefined (pharmacists...!) range of the
prescibed heating rate. You can buy equipment that provides programmed
heating rates.

One problem with USP and EP general methods like this one is there that
there is no traceability - you can't easily find any background information
that explains why a method is defined the way it is, so you feel like an
idiot when you find a problem with it.

Finally, I wonder whether you confuse digital with automatic. There exist
instruments that detect the melting point automatically (generally
optically). They detect when the sample moves, which isn't the USP melting
point, and don't tell you if the compound is melting or decomposing; they
are worth the money in some production environments. That's very upsetting
for quality assurance people as well as analysts, cos the only way to prove
that nobody is cheating on the melting point is to make a - validated -
video of the sample & the display. But here be dangerous territory...

Regards


"information_person" <[Only registered users see links. ]> a crit dans le message de news:
[Only registered users see links. ]...
they
manual


Reply With Quote
  #8  
Old 11-11-2003, 05:22 AM
information_person
Guest
 
Posts: n/a
Default melting point apparatus ?

Chris,

Thanks - i found your insight useful. However your discussion does bring up
a point i didn't formally consider, decomposition.

Wouldn't the literature not distinguish melting point from decomposition and
thus we are really measuring one of the crystal homologs of the compound
melting point and /or decomposition. If the decomposition occurs at a lower
temperature than melting point how is one supposed to determine the melting
point ?

Do you suppose that the manual analog melting points are acceptable as the
digital melting point devices?

And also - what approximate rate of temperature increase should be used for
accurate and reproducable results ? is 0.1 C per minute too slow, how about
1 C per minute or even 5 C per minute?

Thanks.
"Chris R. Lee" <[Only registered users see links. ]> wrote in message
news:[Only registered users see links. ].clara.net...
point,
way
show
quantity
context
to
information
prove
news:


Reply With Quote
  #9  
Old 11-12-2003, 01:01 AM
Bruce Hamilton
Guest
 
Posts: n/a
Default melting point apparatus ?

On Tue, 11 Nov 2003 00:22:35 -0500, "information_person"
<[Only registered users see links. ]> wrote:


Most pharmacopeia methods aren't applicate to materials that decompose
below the melting point, and may also require the sample be dried to
ensure a sharp melting point. Note that the sample is only added as
the expected melting point is approached to minise degradation.


You should get a copy of the pharmacopeia method that's applicable.
For USP 27 ( 2004 ) the section is <741> pages 2324-2325. Note that
USP reuires you to check using one of their standards ( they have 6
reference compounds with specified melting points ).


There are several differing procedures, depending on the compound and
melting point instrument. Basically, you heat the bath or block up to
30C below the expected melting point, remove the thermometer, and
attach the capillary with the sample, put it back in the bath, and
heat at ~3C/min until 3C below the beginning of the melting range,
then reduce heating to achieve about 1-2C/min, and continue heating
until the melting is complete.

For compounds with no method specified you usually heat to 10C below
the expected melting point, but rising at 1.0+-0.5C/min and add the
sample when it's about 5C below the beginning of the melting range.
Youd continue heating until melting is complete.

The first appearance of melting or slumping in the tube is defined as
the beginning of melting, the temperature when the sample is liquid
throughout is called the melting point, and the difference is called
the melting range.

Bruce Hamilton
Reply With Quote
Reply

Tags
apparatus , melting , point


Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On

Forum Jump

Similar Threads
Thread Thread Starter Forum Replies Last Post
What is Gravity? sdr@sdrodrian.com Physics Forum 1 01-27-2008 02:53 PM
Simply put, MOVING DIMENSIONS THEORY is THE NEW MODEL: http://physicsmathforums.com drelliot@gmail.com Physics Forum 0 08-29-2006 06:32 AM
The Achilles Heel of String Theory. S D Rodrian Physics Forum 7 07-08-2006 02:40 PM
GNU units and units.dat; Units of Measurement and Unit Conversion James Redford Physics Forum 0 07-31-2005 12:08 PM


All times are GMT. The time now is 10:25 PM.


Powered by vBulletin® Version 3.8.4
Copyright ©2000 - 2014, Jelsoft Enterprises Ltd.
Copyright 2005 - 2012 Molecular Station | All Rights Reserved
Page generated in 0.22983 seconds with 16 queries